All posts

Texno Medicals

Interested in buying pre owned MyotonPRO device

Texno Medicals  posted 1 days ago in Marketplace
Hello everyone!
We are interested in buying pre owned MyotonPRO Device.
Please, kindly email: texnomedicals@gmail.com if interested.
Many thanks,
Suhrob.
0 comments
JIANBIN
Lefevre

Selling my Myoton device

Lefevre  posted 13 days ago in Marketplace
Hello dear Myoton´s Community, hope all is runing good for you -) I'm selling my MyotonPRO palpation device, which I've used for a clinical study. I'm truly satisfied with its performance and am not selling it due to any issues with the device itself, but rather due to personal circumstances that require me to sell the equipment I own. I highly recommended the Myoton device to the scientific community in Spain, France where i am based. The price is €3,500, with shipping costs to be covered by the buyer. The device is in perfect condition.


Thanks and Regards,
Fabrice Lefèvre
3 comments
Feyer

Price

Feyer  posted 22 days ago in Q&A
Hello,
what is the price for die MyotonPRO?
Thank you and best regards.
Christoph
1 comments
Liis Peipsi
Team

Measurement depth

Liis Peipsi  posted 22 days ago in Did you know

Every measurement location has its unique depth of measurement depending on the oscillating tissue's elasticity and the dissipation of oscillation frequency. The higher the elasticity and more gradual the dissipation of mechanical energy of the oscillation, the further the impulse wave spreads, meaning the greater the depth of measurement.

Average depth of measurement for adipose tissue is estimated to be around 20mm. For skeletal muscles and tendons, depth of measurement is noticeably greater as the impulse wave spreads further due to the dissipation of tissue oscillation being more gradual.



As for skeletal muscles underneath a thick layer of adipose tissue, as a first measure, it is recommended to choose muscles with the least amount of adipose tissue covering them (if possible) to ensure the impulse reaches the muscle. However, when in doubt, measurement depth can be verified by conducting a test measurement with the muscle of interest at full rest vs slightly to moderately contracted. For this purpose, a test measurement with a single impulse can be conducted for both states of muscle activation.
If the second measurement with the muscle contracted reflects higher oscillation frequency and stiffness than the first measurement (e.g. F 14Hz vs 16Hz; S 250N/m vs 300N/m), then the higher values for oscillation frequency and stiffness are due to muscle activation - thus confirming the measurement impulse reaches the muscle and subsequently the muscle response to the measurement impulse reaches back to the device probe.

0 comments
Liis Peipsi
Team

Reference values

Liis Peipsi  posted 22 days ago in Did you know
Reference values have yet to be developed.
Researchers collect their own norms, such as individual norms or base levels for further comparisons. Using the device regularly allows for gaining such experience quickly. Depending on the research being conducted, healthy controls, placebo groups or the unaffected side in patients presenting with a unilateral condition can serve as a reference.


Typically, most superficial skeletal muscles of a healthy adult with the muscle of interest at full rest range from:
Oscillation frequency: 13...15 Hz;Stiffness: 200...350 N/m (m. tibialis anterior usually presents with above average stiffness).

Quantitative values of measurement results are influenced by the following factors:
Muscle activation level (measure at full rest only as it allows for repeatability, the exact level of muscle contraction is not reliably repeatable);Measurement point location (above muscle belly);Muscle length (agonist and antagonist in balance, avoid full extension);Body position (always prefer lying (if possible), this allows for full relaxation and easy repeatability);Physical condition, general health status, medical condition, lifestyle, sports, occupation;Fatigue and recovery status at the given moment;BMI and adipose layer thickness;Age, gender, blood pressure, stress level, temperature.It is crucially important to always use the same conditions and a standardised setup when collecting measurements.


The development of reference values is a challenging task but can certainly be accomplished. The project would require international scientific collaboration to establish and describe classification criteria. A large multi-centre study is necessary with data collected from diverse populations.
0 comments
Liis Peipsi
Team

Tap time / Impulse time

Liis Peipsi  posted 22 days ago in Did you know

The default impulse time of 15 ms is suitable for all tissue structures with a thickness above 5 mm and a stiffness below 500 N/m. A decrease in impulse time is necessary for stiffer less compliant tissue structures (i.e. tendons, ligaments).

With the aim of improving tissue response to the measurement impulse, the latest model MyotonPRO 1B automatically reduces impulse time for stiffer tissue structures as follows:
15 ms, standard impulse time – Suitable for tissue structures such as adipose tissue and skeletal muscles with a Stiffness (S) below 500 N/m;13 ms – S 500 N/m…700 N/m;10 ms – S 700 N/m…1400 N/m;7 ms – S 1400 N/m…2000 N/m.



The previous MyotonPRO model allows for manual reduction of impulse time (also known as Tap time). The above mentioned values should be taken into consideration.


0 comments
Aleko Peipsi
Team

Guidelines for Dual-Hand Operation of the MyotonPRO Device

Aleko Peipsi  posted 1 days ago in Did you know
For precise and efficient use of the MyotonPRO device, a dual-hand operation technique is recommended. This approach improves control over the device's measurement position, enhancing both stability and convenience of measurements.
The measurement position's axial range is 3.0mm wide, necessitating good control. To facilitate this, gently hold the device with your dominant hand and perform micro-adjustments using your wrist. Simultaneously support and stabilize your dominant arm using your other hand, ensuring that the device remains steady within the required range.


For accurate positioning, the MyotonPRO device is equipped with three indicators:
1. Green Probe Light: Indicates when you are entering or leaving the measurement position.
2. Probe Marker: A 3mm-wide marker on the probe serves as a visual guide to the measurement postion.
3. On-screen Instructions: Clear instructions such as 'Move into position!', 'Hold position!', and 'Move back!' to assist in the measurement process.
While conducting measurements, it is important to keep focus on contact between the probe and the skin to avoid any lateral movement. Simultaneously, make use of the green light and the probe marker as guides for axial position of the device. This approach not only helps in maintaining proper contact between the probe and the skin but also facilitates convenient axial control of the measurement position.


Note: For optimal results, aim to position the device so that the probe marker is halfway exposed at the probe opening, indicating the center of the measurement range.
Employing the described technique ensures effective and precise operation of the MyotonPRO device, contributing to the utmost reliability of the results.
The accompanying images illustrate this instruction.



0 comments
Aleko Peipsi
Team

Creating a New Database in Myoton Software

Aleko Peipsi  posted 5 days ago in Did you know
Creating a new database in Myoton software is a straightforward process! Follow these simple steps, as illustrated in the accompanying screenshot:
1. Open Myoton Software: Launch the application on your computer.
2. Open Connected Database Location: Click on the Database icon located on the main toolbar.
3. Rename Connected Database: Modify the existing database name to suit your project and click OK. These steps allow you to create as many databases as needed, for example, one for each project.



Database Name: Choose a name that reflects your project title. It allows you to keep measurement results from different projects separate, enhancing data management.
Flexibility and Efficiency: You can easily switch between databases as needed. Additionally, the Myoton software offers the capability to merge or combine multiple databases. This feature is particularly beneficial for multi-center studies, allowing you to consolidate sub-databases from various sites into one comprehensive database. This step may be beneficial before conducting final data analysis and statistics.


By following these guidelines, you can effectively organize your data in the Myoton software, enhancing your workflow, allowing the software to operate with smaller database sizes. Moreover, it is also safer to keep different projects in separate database files.
2 comments
Alastair McLoughlin

Menstrual cycle and Myoton measurements

Alastair McLoughlin  posted 7 days ago in Q&A
Has anyone noticed changes in tissue tension or relaxation of the tissues being measured in women during their period of ovulatory or menstrual phases?
Would measurements using the Myoton reflect changes in projected readings due to oedema (fluid retention) during these phases of a woman's cycle?
Has this even been measured or factored in to results- especially on abdominal tissues?
The link to some research I found, shows that strength is affected due to higher levels of oestrogen at mid-cycle.
https://www.ncbi.nlm.nih.gov/pmc/articles/PMC1158967/pdf/jphysiol00289-0263.pdf
Have the effects of the hormone 'Relaxin' been factored into anyone's research results conducted on women, just after ovulation? Would Relaxin noticeably affect Myoton measurements?
Thanks for any comments.


2 comments
Aleko Peipsi
Team

Standardization aspects

Aleko Peipsi  posted 10 days ago in Did you know
Standardization of the measurement conditions is important for maximizing the reliability and repeatability of the results.
Here are some aspects to follow:
Skeletal muscles and tendons should be measured at full rest (EMG signal silent);Preferably, the subject should be in a lying position to ensure full relaxation and consistent repeatability between sessions.Schedule measurement sessions before, rather than after, any interventions or physical exercise.Conduct measurements at room temperature for optimal results;Ensure consistent identification of anatomical measurement points in interest;Avoid physical contact with the tissue being measured (except the device probe);Avoid distractions during the measurement;Avoid tight clothing which may influence results;Minimize direct contact with the subject, allowing only necessary contact for operator and device stabilization;Make sure the device probe is held perpendicular to the skin surface;Ensure the device is gently held with both hands for optimal stabilization.
0 comments
Alastair McLoughlin

C-section scar results

Alastair McLoughlin  posted 11 days ago in Q&A
We have just concluded a feasibility study into the effects of treatment for C-section scars.
If I may impose on your time, I am having just a little difficulty interpreting the results, and would like your expertise to assist me.
From the PDF file I have attached, it seems that all 5 parameters of measurement show changes in the scar tissue. 
Frequency Hz, Stiffness N/m and Relaxation ms show improvement but Decrement (plasticity) and Creep have increased. If I am correct this means that although the tissue is more relaxed it keeps it's shape better when being stretched.
Forgive me for my lack of understanding, but just seek some guidance from you to understand what these numbers signify.
With appreciation.
Sincerely,
Alastair
1 comments
Aleko Peipsi
Team

Join the Myoton Community Forum!

Aleko Peipsi  posted 17 days ago in Support
Dear MyotonPRO user!
Exciting news for all MyotonPRO palpation device users and researchers! Myoton has launched a brand-new community forum dedicated to connecting MyotonPRO device users worldwide. Whether you're an experienced user, an academic or clinical researcher, a medical practitioner, or even just considering the acquisition of the MyotonPRO device for your new research on soft biological tissues, this forum is your go-to platform.
Direct communication between users helps to increase awareness about the versatility of the palpation device, build up knowledge about its wide range of applications, learn about results interpretation, and consider its practicality as the emerging new standard for soft tissue assessments.


Why Join the Myoton Community Forum?
Share Experiences: From breakthrough discoveries to minor challenges, share your experience and learn from others.
Ask Questions: Unsure about specific applications, features, or limitations, or want some tips on how to achieve the highest reliability, or how best to prepare efficient measurement collection? The community is here to help.
Provide Support: Are you an expert user with extensive experience and in-depth knowledge of the device or software? Help fellow users by answering their questions, offering guidance, and providing solutions.
Stay Updated: Be the first to know about news, updates, new features, or offers related to the MyotonPRO device.
Direct Interaction with the Myoton Team:
Have specific questions or concerns? The Myoton team is actively present in the forum, ready to assist and provide insights. Your queries will not only get authoritative answers but will also contribute to building a comprehensive knowledge base for the community. Over time, this will evolve into an invaluable FAQ, making the Myoton experience smoother and more informed for everyone.
The Myoton team eagerly anticipates user feedback and fresh ideas regarding potential enhancements to our device or software. We're keen to understand which aspects you appreciate and which areas you believe require improvement. Your insights on the functionalities and suggestions for enhancing or simplifying the user interfaces are extremely valuable and serve as crucial input for further product development.
Building a strong, supportive community is essential for all of us — users, clients, and product developers alike. With the Myoton Community Forum, users can not only troubleshoot and learn but also celebrate each other's successes and milestones.


Join to the Community!
If you're passionate about the Myoton palpation solution or simply curious to learn more, there's no better time to join the conversation. Together, let's make this forum an invaluable resource for all.


Sincerely,
Aleko Peipsi
0 comments
Top contributors
Aleko Peipsi with 9 posts and comments
Team
CEO in Myoton AS
Liis Peipsi with 3 posts and comments
Team
Alastair McLoughlin with 3 posts and comments
Lefevre with 2 posts and comments
Daniel with 2 posts and comments
Feyer with 1 posts and comments
MIJIN KIM with 1 posts and comments